11-16-2015 11:49 PM - edited 11-16-2015 11:51 PM
I want to measure the changes of current on the pressure indicator. For this purpose I use module SCC-CI20 in my sc-2345 connected to the computer with NI DAQPad-6015.
In MAX all the devices are displayed correctly, but, when I start the process of measuring there are only strong scattered interference noises on the display, and it seems like the signal from the pressure indicator doesn’t go through DAQPad-6015 .
I think that the problem is that NI DAQPad-6015 cannot identify or see the signal on sc-2345, but I don’t know what to do to solve this problem.
11-18-2015 10:53 AM
Hi Leonid.Rybalkin,
What range of currents are you inputting from your transducer? Also, are you doing your measurements through MAX?
11-18-2015 12:09 PM
yep
When the circuit is closed, I connect the tester to the module, and everything is displayed correctly. Here are the pictures of the work.
1 image : DC 15V 100mA - off
2 image : DC - on
11-19-2015
06:42 PM
- last edited on
03-14-2025
04:52 PM
by
Content Cleaner
I took a look at the manual for the SCC C120 (https://www.ni.com/docs/en-US/bundle/371068b/resource/371068b.pdf) and based on page 6, it looks like it has a recommended maxiumum input current of 20 mA. That may be the cause of any unusual behavior