Detect TTL signal using configured acquisition
I am using configuired acquisition to acquire signal in NI PCI 5105 in visual studio 2008. The other signals are fine but when I give a TTL signal input having a low duty cycle, it can not detect the TTL signal. Is it because it has low duty cycle? The signals that are coming only shows the value around 0. I tried to connect the same signal to NI Scope front panel and it is showing the correct signal. What can I do to solve this problem? I've attached a screen capture of scope front panel. Here we can see the trigger signal in channel 1 but this signal can not be seen using configured acquision.cs. Please help..
Message 1 of 30
(6,261 Views)
How low is your duty cylce? Is configured acquisition.cs example code or something you have written? Are you able to trigger if you increase your duty cycle?
Rohama K.
Message 2 of 30
(6,255 Views)
The duty cycle is as low as 0.4% I have used configured acquisition.cs that NI has provided and made some changes in it. I even tried with the unchanged code but it is not showing the TTL signal. I used the 1 KHz sinusoidal signal in the same channel and its working fine. Please help me with this.
Message 3 of 30
(6,251 Views)
Hi Paritosh,
It seems your duty cycle is very low. The duty cycle tolerance for the PCI-5105 is 45% to 55%. Specifications on the duty cycle are listed on page 11 of the PCI-5105 Manual. Does that help to answer your question?
Travis Ann
Customer Education Product Marketing Manager
Message 4 of 30
(6,237 Views)
But I am getting the signal in Scope Soft Front panel as you can see in the attachement in my previous post. Also, I used the same TTL signal as a trigger in a labview program. I need this signal in channel 1 to act as a trigger for channel 0 source.
Message 5 of 30
(6,233 Views)
Hi Paritosh,
Can you clarify for me how you have you triggering configured? I see that in your SCOPE Soft Front Panel shows that you are using rising edge trigger CH 0 (sine wave) to trigger CH 1 (TTL), but you describe that your code is trying to trigger off of CH 1 (TTL) to trigger CH 0 (sine wave).
Also, what is the frequency and voltage range of these signals?
Travis Ann
Message 6 of 30
(6,213 Views)
The trigger in the attached picture is just to test. The real trigger will be the TTL signal in channel 1. The TTL signal is a rising edge trigger in channel 1. The frequency is maximum of 1 KHz but it sometimes can go down to 200 Hz or 500 Hz as well. The voltage is around 4 volt and the trigger level is around 3.5 V.
Message 7 of 30
(6,208 Views)
Hi Paritosh,
From what you are describing, it sounds like what you have set up should work. Could you post some screenshots of your set up so we could take a deeper look?
Travis Ann
Message 8 of 30
(6,196 Views)
Sorry for replying late..Actually I managed to get the TTL signal as a trigger as you can see in the attached image. However, the data acquiring rate is too slow although the data is coming to the digitizer at 1 KHz. I am taking 500 records for each signal.  How can I speed up the acquisition time? Please help me on this..
Message 9 of 30
(6,183 Views)
Hi Paritosh,
The acquiring sample rate is set in the Configure Horizontal Time function in the NiScope driver, so if you want to acquire at a faster rate, you will implement that in your code. You are limited to the overall acquisition rate of the card, which in your case is 60MS/s.
Travis Ann
Message 10 of 30
(6,173 Views)