10 V Instrument hooked to 8234

We have a Kistler Charge Amplifier which puts out +- 10V hooked to a 9234.(only _-5V). I was told what would happen would be just clipping at 5V.

This seems not to be the case. When th 9234 is fed to ME Scope, and I convert there Volts to Units, I get the equivalent of 50 V registered. No sign of the clipping, signal seems to behave like everything is fine, What is this digitizer doing? ME Scope just gets a voltage valune, it should reach a max of +-5V, certainly not 30V! Any idea what is happening.

0 Kudos
Message 1 of 2
(801 Views)

Personally, I would never connect a sensor capable of +/-10V to a +/-5V capable instrument unless I am 110% sure that it will never exceed +/-5V.

 

Few questions,

  • Where did you get the info about the clipping or clamping behaviour?
  • What is a ME scope?
  • Could you please share the sensor model, any additional details about your code module or VI implementation settings, scope capture etc.,
  • Please share some sampled data using 9234 for analysis

 

What I read from the datasheet is that the overvoltage protection kicks in only at 30V and not at 5V

santo_13_0-1629173445477.png

 

santo_13_1-1629173527247.png

 

 

Santhosh
Soliton Technologies

New to the forum? Please read community guidelines and how to ask smart questions

Only two ways to appreciate someone who spent their free time to reply/answer your question - give them Kudos or mark their reply as the answer/solution.