Hi Good day, I am try to find an answer about on how to connect my 3-wired pressure transducer to an NI 9219 universal analog. NI 9219 is the only module that is available in our institution and only module that is compatible in reading pressure transducer. I'll be honored and thankful if someone can fixed my problem on how to connect my pressure transducer to NI 9219.
07-14-2023 08:30 AM
Can you share more details on your Pressure Transducer? model? link to a datasheet or user manual?
07-14-2023 08:36 AM
Attached photo is the pressure transducer that I bought.
Here are the Specification:
Item Type: Pressure Transducer
Material: Stainless steel
Overload Pressure Range: 100PSI
Overload Pressure: 1.5 times rated pressure
Breaking Pressure: 3 times rated pressure
Operating Temperature: -40℃-125℃
Output Signal: 0.5V-4.5V linear voltage output
Accuracy: Within 2% of reading (full scale)
Thread: 1/8in 27NPT thread
Media Compatibility: Refrigerant, clean non corrosive oil, gas and water, etc.
Wiring Connectors: Watertight quick disconnect. Includes mating connector (standard Packard Metri pack connector) and wiring harness (pigtail).
Wiring: Red +5V. Black for ground. (sometimes green) Blue for signal output
07-20-2023 05:50 AM
Your sensor needs a 5V supply voltage , as far as I understood the 9219 spec ,this unit can't supply 5V
so you need an external 5V supply (red and black) . the voltage out pin (blue) and black go to the 9219
07-25-2023 03:40 AM
Can you please elaborate further? i didnt get it totally especially the the color assignments
Thank you
07-25-2023 04:20 AM - edited 07-25-2023 04:28 AM
What cable colors do you have on your sensor?
Since you already ask aboout the cable colours, I assume you have no engineering background.
You need a 5V power supply. You can try one of these 500mA 5V chargers, but you need to check if they really put out 5V (some do more) and usually the 5V provided is not very 'clean' , mean you migth measure some nasty hum or other stuff due to the switched power supply. Nice would be a bench lab powersupply set to 5V.
the black(or green) cable from your sensor needs to be connected to the negative powersupply and the [negative/GND/AI- ] input of your 9219 channel . the red cable need to connected to the +5V of yur powersupply and the blue to the [positive/AI+] input of the 9219.
07-25-2023 05:04 AM
Thank you for the reply.
My sensor has Green, Black, and Red wires.
08-06-2023 08:51 PM
Greetings,
the sensor still wont read. I've the wire connection you've suggested
08-06-2023 09:21 PM
@Jeiel97 wrote:
Greetings,
the sensor still wont read. I've the wire connection you've suggested
Unfortunately, there is no way we could verify your connections and how you attempted to read the sensor.
Please provide all relevant information as we are limited to what you post here.