SuperWolfi,
If you look at the 6025E User Manual, in the Specifications Appendix A, you will see that for a gain setting of 1 you will have an input range of +/-5V. If you use a gain setting of 0.5 then your input range is expanded to +/- 10V.
In your call to AI_VRead you are using a gain of 1, instead of 0.5. For information on how to set the gain for your board, please reference Appendix B:
Analog Input Channel, Gain Settings, and Voltage Calculation of the NI-DAQ Function Reference Manual. You will find that -1 translates to a gain of 0.5.