self-calibration of NI PCI-6036E
I have a new NI PCI-6036E (low cost) data acquisition board. We set it up using one of the analog input channels in single-ended nonreferenced mode.
I want to acquire a voltage value which is equal to a pressure reading of the connected pressure controller.
After connecting all parts together and set up the parameters I still have an offset between the voltage value from the pressure controller (7.85 V) and the card reading (8.05 V). How can I solve this problem?
I tried to do a self-calibration of the data acquisition board using the provided online E series diagnostic utility on your web page undere: www.ni.com/support/selftest but it says "device detected but not supported". The E-series model I am using sould be supported
Can I do a quick self-calibration in an other way?
It would be very nice somebody could give me some advice how to solve this problem.
Thank you very much for your help!
beam
(2,735 Views)
Beam,
You can programmatically perform a self-calibration. In LabVIEW, you can call the E-Series Calibrate VI within the Data Acquisition>>Calibration and Configuration subpalette. If using a text-based ADE, you can call the Calibrate_E_Series NI-DAQ function.
You are right in that the PCI-6036E device should be supported by the online diagnostic utility; that should work. Make sure you have NI-DAQ 6.9 or higher and Internet Explorer 5.0 or higher.
Also, is the voltage you are reading floating or ground-referenced? If the source is floating then you should use bias resistors when configured in NRSE or differential mode. Chapter 4 of the E Series User Manual (linked below) goes over how/when to connect bias resistors. I would take a look at this chapte
r and use it as a reference for connecting signals. I hope this helps.
PCI E Series User Manual
http://digital.ni.com/manuals.nsf/webAdvsearch/06F1C9FB0D0BA5C286256C010057461B?OpenDocument&vid=niwc&node=132100_US
Todd D.
NI Applications Engineer