04-08-2009 07:54 PM
Hi All-
I'm brand new to labview and DAQ in general, so sorry for the newbie question. I'm using a 9014 on a 9101 chassis with a 9237 module and a full bridge load cell, trying to run the 9237 getting started example. The problem is that the ouput from the module is stuck at 0.025011v. I don't think it's a problem with the bridge, I checked it with a standalone strain indicator and it works fine. Any ideas?
Thanks
-Drew
04-09-2009 10:20 AM
I have a few questions to help narrow down the issue.
Is it a problem with all channels on the 9237 or just one of them?
Does it happen in all slots of the 9101 or just one of them?
Where are you reading in this voltage, in Measurement & Automation Explorer, or in LabVIEW?
Do you have a power supply that you can use to send a constant voltage into the 9237 and then try to read that voltage? (be careful about the max/min voltage of the 9237 listed in the spec)
04-09-2009 02:25 PM
Hi Doug, thanks for the speedy reply-
I have a bridge wired into channel 0, the others are empty. They read -0.024980, -0.02366, and -0.02498 respectively. I'm using labview to read this voltage.
I just tried moving the module to slot 2- channel 0 still reads the same, but the other channels are all over the place now. Can i disregard these channels when there's no bridge attachced or does this indicate a problem?
I'll see if I can dig up a power supply to hook up to the 9237.
Thanks
-Drew
04-09-2009 02:29 PM
Also- does the 9237 require the shunt calibration and remote sense terminals to be connected, or can i disconnect them for simplicity?
Thanks
-Drew
04-10-2009 09:58 AM
It sounds like the voltage is railing to the ADC because the max voltage read in by the card is usually 25 mV/V. What kind of bridge is this (full bridge, half bridge, quarter bridge) and how is it wired up the pins (include excitation please). Also what excitation level are you using and what is the nominal resistance of your bridge, as there is a maximum supply power that the 9237 can supply.
The shunt resistor does not need to be connected unless you are performing shunt callibration. The remote sense is also not required but can be used to ensure accuracte excitation voltage supply.
04-12-2009 06:18 PM
It's a full bridge, wired as follows:
Pins 1, 10: shunt calibration
Pins 4, 5: RS+ and RS-
Pin 2: AI+
Pin 3: AI-
Pin 6: Ex+
Pin 7: Ex-
The gages are 350 ohm, and I left the excitation voltage at 3.3v.
Thanks
-Drew
04-13-2009 11:51 AM
04-13-2009 04:22 PM
04-24-2009 04:13 PM
Hey everyone,
Drew called in and we figured this one out on the phone. Turned out to be a bad solder joint on his connections.
For future reference: