Hi SirMutt,
This can occur if you are set up taking RSE measurements and your external circuit is at a different ground than your DAQ device. This can create a ground loop and thus, the measured difference. I would recommend opening a test panel in MAX, setting the input configuration to Differential and connecting the positive voltage to ai0 and the other end of the resistor to ai8. If this is not an option, you can also try an input configuration of NRSE and connect the negative terminal to AISENSE.
The 6251 does have a large input impedance and should not be loading your circuit. Are the other instruments connected at the time or are the connections across the resistor made independently? I hope these ideas help and I recommend looking at the "Field Wiring and Noise Considerations for Analog Signals" article in our developer’s zone for some reference. In other news, your icon and username crack me up! Is that the dog from Duck Hunt?
Field Wiring and Noise Considerations for Analog Signals