01-14-2008 04:10 AM
01-14-2008 10:26 AM
To give some more information about my problem;
The pressure transducer has 3 connection points on its backside. Let's call them connection point (1), (2) and (3). To connection point (1) and (2) are a ~30V power supply connected. To connection point (2) and (3) are the output cables connected. With these output cables a 4-20 mA signal can be read with a standard multimeter. I am still not able to get any reading using the NI 9203.
It says in the manual of the pressure transducer that only an ammeter whose internal resistance less than 10 ohms should be used. Can this be the source of my problem?
01-14-2008 11:09 AM
I have tried something additional, which again did NOT solve my problem. I had a NI 9205 (measures voltage) laying around, and I connected the 4-20 mA signal from the pressure transducer to the NI 9205 with a 470 ohm resistor in parallell. I have done this many times before, and never had a problem. Again, the same type of reading appeared as when I was using the NI 9203 (measures current).
Anyone has any ideas?I
01-14-2008 01:52 PM
Hi EspenSH,
@EspenSH wrote:It says in the manual of the pressure transducer that only an ammeter whose internal resistance less than 10 ohms should be used. Can this be the source of my problem?
01-15-2008 10:40 AM
I have managed to get some sort of reading using the NI 9203. The mA signal from the pressure transducer should vary between 4-20 mA. What I read in Labview or MAX is a signal from 3.2mA - about 7mA.
Some further information: the output potential from the pressure transducer at 4 mA is around 0.66V, at ~12 mA around 0.72V, and at ~20 mA 0.78V. By measuring the current with the multimeter only gives a nice 4-20 mA reading.
I tried to connect a 9205 unit in parallell with both a 470 ohm resistor as described above, and also resistors of ~4 ohms, ~ 10 ohms and ~20 ohms, but without any more success.
I would really like to see the 4-20 mA signal in Labview, since the pressure transducer is calibrated from manufactor with respect to the 4-20 mA. (I need to measure accurate pressures). If it is the case that the internal resistor in 9203 is too large (manual of the pressure transducer indicated using an ammeter with max 10 ohms) what can I do in order to still use the 9203 and get a 4-20 mA signal in Labview?
01-21-2008 10:15 AM
01-21-2008 10:35 AM
06-28-2018 12:21 PM
Hello,
I have issue to measure some current with a transducer Keller. Whatevver the pressure applied, the currant is very very low.
I have tested :
1- direct connection to the NI9203 (current 1.10e-7)
2- connection on shunt resistance of the the NI9203 0,0018A
I think understand that the transducer and NI9203 have to be wired in serie with the 24V. Is it correct?
the transducer is conected between 1 &3.
Thanks for your help.
Sebastien
06-28-2018 12:22 PM
nota: my DAQ is cDAQ 9185