Can the NI6229 analog input channels output anything?
I know the answer is probably no, but I'm asking anyway because if there is a way to have the analog input channels output a voltage it would make selftest simpler.
(3,290 Views)
No, they can't output anything, but they can measure internal channels, such as the board's analog outputs and calibration reference if that helps with your self-test.
Chris
(3,263 Views)
Hello Steve,
You are correct; the analog input channels on the NI 6229 cannot output a voltage.  However it sounds like what you're trying to do is a self test.  If you are wanting to route the 6229's analog output channels to an analog input channel to read back the output then this is something that you can do internally.
If you just want to test the card you can use the DAQ Diagnostic Utility.  This utility internally routes signals to make sure the card is functioning correctly.
If you want to view the output then you can use the test panels in the Measurement and Automation Explorer to do so.
1.  Open the Measurement and Automation Explorer.
2. Navigate to My System» Devices and Interfaces» NI-DAQmx Devices.
3.  Right click on the PCI-6229 and select Test Panels.
4.  Click on the analog output tab, configure a sine wave output on ao0 and click start.
5.  Go to the analog input tab and select DevX\_ao0_vs_aognd, set the sampling mode to continuous, the rate to be large enough to sample your sine wave and the samples to read to be 10% of the rate.
This will allow you to read back the analog output values without connecting any wires externally.  I hope this helps, and if this isn't what you're looking to do it may be helpful if you can provide some more information about what you're trying to do.
Cheers,
Brooks
(3,262 Views)
I wanted to test the presence connections in my test box without having to wire or jumper anything else. I was wondering if there was a way to internally route a signal to the pin so the pin could temporarily be used as an output.
(3,255 Views)
Hello Steve,
In order to do this you will need to change your wiring externally.  While you can route signals internally this doesn't make them available on the external pins.
I hope this helps and have a great weekend!
Brooks
(3,228 Views)