Hi Nic,
Unfortunately, what Dennis has said is correct. (Well, it’s good that what somebody posts is correct, but it’s unfortunate that it’s not the answer you were hoping for.) The only two hardware solutions are:
1. Get a device that has the ability to sample as quickly as you want, or
2. Get a second device (one for faster sampling, one for lower sampling)
The option Dennis mentioned would make your acquisition software-dependent, so don’t expect consistent timing between samples, but you could choose to do that if you’re aware of the consequences.
I did a search on ni.com/daq to
narrow down USB devices with more than 16 channels (32 is the next step up) and
100kS/s. For 13kS/s for all 22 channels, you would need a device that can
sample at least 286KS/s, which would leave the USB-6289 (625 kS/s), USB-6259
(1.25 MS/s), or the USB-6255 (1.25 MS/s). They have different specifications
for resolution, number of AO/DIO channels, etc., and if you want to ask more
specific questions about each one, feel free to visit ni.com/contact to speak
to a Sales representative.
I hope that helps. Best of luck with
your application.