DAQ USB-6451 (OEM) : Analog Input (AI) peak speed
I’m working on capturing 16 analog input (AI) channels.
The USB-6451 can acquire 8 AI channels simultaneously at 100 kS/s, so I configured it for 50 kS/s when using 16 channels.
According to the manual, channels share ADCs in pairs — for example:
  • AI1 and AI9 share one ADC
  • AI8 and AI16 share one ADC, and so on.
Issue
When the sample rate is greater than 200 kS/s (for example, 500 kS/s) and I record about 100 samples, I see a problem if paired channels are at different voltage levels.
For instance, if AI6 = 4 V and AI14 = 3 V, then during the first few samples:
  • AI6 shows a voltage drop, and
  • AI14 shows a voltage rise.
All 16 channels are configured as Reference Single-Ended (RSE).
Observation
  • The problem disappears when the sample rate is below 200 kS/s.
  • It also disappears when both channels are at the same voltage level.
(173 Views)
Accepted by topic author Coherent_L
You may be experiencing Ghosting in multiplexed ADC architecture DAQ
Please read,
How Do I Eliminate Ghosting from My Measurements? - NI
Santhosh
Soliton Technologies
(146 Views)
Thank you for your solution!
Just wanted to confirm my following understanding:
NI explicitly notes that dummy channels count toward the device’s maximum aggregate sampling rate.
If I currently scan 16 real channels and then add 8 dummy channels (one per ADC pair), the scan list becomes 24 channels.
As a result, adding 8 dummy channels will reduce the maximum per-channel rate from 500 kS/s to approximately 330 kS/s, unless I increase the device’s allowed conversion time (interchannel delay).
(115 Views)
Or you can lower the output impedance of your signal to lower the settling time and minimize the ghosting effect.
Eliminate Ghosting on Adjacent Input Channels by Decreasing Source Impedance - NI
What signals are you connecting to the AI?
Santhosh
(95 Views)
we are working on customized Optical power meter, we’re connecting a photodiode (PD) PCBA to the NI USB-6451, the signal on each AI (analog input) channel is almost certainly a analog voltage
(74 Views)
So, the PhotoDiode doesn't have any amplifier or buffer in the path? if not, you must include one before connecting to the DAQ, as the output impedance of a PD will be typically high and be easily affected by ghosting.
In general, lower the output impedance of these Analog signals, the effect of ghosting will be minimal.
Santhosh
(69 Views)
Thank you for your reply, we already reduce it and small enough, I guess that is our limitation. But anyways we will not take those unstable samples as my solution.
(18 Views)
