Hi Victr,
I would make sure that each channel set up has the range of 10 V to -10V. The default setting in MAX is 5 to -5 V and if you added the channels before making the voltage adjustment, some of the ranges might still be set to a max of 5 V.
Some questions:
Can I ask what the linear scale you are using is? If you remove every channel but one from the task and then apply the scale, do you see the same problem?
Is it all five channels that rail at 5 V or only one?
Also, how are you saving your data to an lvm file? Are you running the task in MAX or using the task constant to run it in LabVIEW and then save the data to a file?