02-22-2010 07:04 AM - edited 02-22-2010 07:06 AM
Hallo,
i`m using PXI-6255 with 80SE / 40Diff for my measurementschane. One of it was the voltagedrop of a +14V Cable:
I was sure that i can use my Channels in Diff.-Mode only with Inputvoltage +/-10V (http://www.ni.com/pdf/manuals/371291h.pdf Page 1) ! That means, i`m using bleeder to divide my voltage!
I was searching in the forum and i read this answer in a thread:
Each input is limited to +/-10V from ground in DIFF mode, so:
(-10V) - (10V) = -20V
(10V) - (-10V) = +20V
Thus +/- 20V range. Unless I am totally missing the boat here.
I looked up the manual, it tends to agree (portions copied/pasted here):
http://www.ni.com/pdf/manuals/371303g.pdf
The differential input mode can measure ±20 V signals in the ±20 V range.
However, the maximum voltage on any one pin is ±10 V with respect to
GND. For example, if AI 1 is +10 V and AI 5 is –10 V, then the
measurement returned from the device is +20 V.
Input range
Single-ended ................................... ±10 V
Differential...................................... ±20 V, ±10 V, ±5 V, ±4 V,
±2.5 V, ±2 V, ±1.25 V, ±1 V
Working voltage..................................... ±10 V
Now i`m confused, can i use my AI+ /AI- directly without a bleeder (e.g. 14V AI+, 13,8 AI-)??
I tried it once and it works without problem.
Thanks in advantage
G. Jilwan
02-22-2010 12:53 PM
If I understand your question correctly, my answer would be no, you can't. Let me make sure I understand you. You are thinking about putting 14V on AI+ and 13.8V on AI-, and from that configuration you intend to get 0.2V in differential mode. Is that correct? This setting doesn't exceed the differential limit, but exceed the single-ended setting. The spec clearly states that each pin shall not have voltage of greater than +-10V. The voltage 14V and 13.8V are greater than +-10V, so you can't do that.
Yik
02-23-2010 01:59 AM
Correct!
Since i was sure that i cannot use an input range higher than 10V, I was confused reading this message in the forum with a USER GUIDE AND SPECIFICATIONS USB-6008/6009. I didnt see the difference between the 2 devices (6255 and 6008)But it`s ok i`ll not stop using my bleeder!
thanks
regards
02-23-2010 03:25 AM
Depending on the output impedance and signal bandwidth you can use a simple resistor voltage divider to expand the input range.
02-23-2010 06:51 AM
Yap that what i meant with bleeder = Resistor voltage divider!
i`ll stay on this point!
Thanks a lot!
Regards
02-23-2010 08:59 AM