Error -200153 when adjusting PXI-6115

Calibration Executive v3.1.2
 
I'm running a PXI-6115.  While adjusting the module, Cal Exec returns an error:
---
An NI-DAQmx interface error occurred at:
Step: DAQmx External Calibration
Sequence: Adjust E Series
 
Specific Error Information:
Error Code: -200153
Error Message: Reference Voltage applied for calibration is outside the range defined for calibration of this device.
 
Ensure that the reference voltage falls within the range specified for the device.
---
 
I've ran the module 3 times, & I get the same error every time.
 
 
Brian Johnson
PMEL Tech III
Bionetics PMEL
Wright-Patterson AFB
0 Kudos
Message 1 of 6
(3,838 Views)
Hi Brian,

What kind of calibrator, DMM, and counter are you using in your calibration setup?  How far are these devices out of calibration?  Is this the only PXI-6115 that you have and are trying to calibrate, or do others pass calibration successfully?  In addition, are you using revision B of the E Series calibration fixture (S Series devices require this revision)?

Regards,
Andrew W
National Instruments
0 Kudos
Message 2 of 6
(3,826 Views)
Brian,
 
I suspect that what you are seeing is due to the version of a versioning conflict between CalExec and DAQmx.  If you are using DAQmx 8.3.1 or newer, then you need Calibration Executive 3.2 for 6115 devices.  Here's some background on why this is:
 
The actual calibration procedure is implemented in the driver, which CalExec calls into.  We found a way to improve our calibration algorithm for 6115 devices in the driver, and this got implemented in DAQmx version 8.3.1.  However, to implement this change, we needed to change the allowable value of the reference voltage applied during calibration.  We used to allow any voltage between 6 V and 9.998 V.  We changed this to be 5 V +/- 5 mV.  The problem here is that if you are using an older version of CalExec (older than 3.2), it will apply a reference voltage in the old allowable range.  If you have updated DAQmx to 8.3.1 or newer, this is no longer valid.
 
If this is the problem you are seeing, then there is two ways to fix it.  As a first option, you could upgrade CalExec to version 3.2, which was updated to work with the new 6115's new calibration algorithm in DAQmx drivers 8.3.1 and newer.  Your second option would be to back your DAQmx version down to something older than 8.3.1.
 
Hope this helps,
Dan
Message 3 of 6
(3,818 Views)

You hit the nail on the head.  I'm using DAQmx 8.3.1 & Cal Exec 3.1.

Is there a patch I can download to update Cal Exec to 3.2?  If so, could you provide a link?

 

Thanks

0 Kudos
Message 4 of 6
(3,794 Views)
Brian,

Calibration Executive 3.2 is a paid upgrade version because it is a full released application.  If you want more information about purchasing this version you could contact a sales representative at (888) 280-7645.  If this is not an option, downgrading the installed version of NI-DAQmx will be required.

Regards,
Andrew
0 Kudos
Message 5 of 6
(3,784 Views)

I don't think upgrading the software will be a problem.  The AF Research Lab has deep pockets.  😉

Thanks for the info.

0 Kudos
Message 6 of 6
(3,780 Views)