Floating Voltage Problem
We have recently purchased an NI PXI system with an X series DAQ card (6363). PXI system is controlled by a laptop via an express card (no onboard computer).
I have some troubles measuring the analog voltage output of a torque sensor.
The torque sensor has five connections:
1. Supply
2. Supply GND
3. Torque Voltage Output
4. Torque Voltage Output GND
5. Shield
The problem is as follows.
When I measure the voltage between lines 3 and 4 using a voltmeter, I read a constant value of 0.051 V. When I connect the same lines to the DAQ device, the value floats between 0.02 V and 0.08V if RSE is used.
I have also tried differential setup. I connected AI - to AIGND. I have also tried connecting resistors ( 10 K to 100K) between AI-,+ and AIGND. Although some improvement is observed, the values are still floating. The strange thing is that the value read from the voltmeter is always constant.
The vi is simple. I have just used DAQ assistant. Single channel. Screenshot is attached.
I was using a PLC before which also provides non-floating value for this sensor.
Any help would be greatly appreciated.
(3,614 Views)
Could that 'floating' be a result of line noise (Hum) sampled at 100Hz ?
If you sample 1000 points at 1kS/s (or 10kS/s), how does your signal look like?
(If you have an analog scope, how does it look like?)
If you do a AC measurement with your voltmeter, what does it show (60mV??)?
How fast is the signal of your torque-sensor changing?
Your stable signals are likely a result of a lowpass filter < 50Hz ;).
Greetings from Germany
Henrik
LV since v3.1
(3,606 Views)