Hello,
This is just to notify others that we are allready working on this one. As we disscussed on the phone this morning the problem is most likely occuring due to bias current from the module. It looks like some load is building up because there is no return path to ground for the BIAS current. I advise to try a bias resistor from the negative lead of your thermocouples connected to the COM pin of the module. A second option you can try to see if BIAS current is indeed the problem is connecting the thermocouples in Single Ended mode.
Hope this helps, let me know if there are any other issues.
Regards,
RikP - National Instruments Applications Engineering