GE Humitrac 4-20mA Output NI 9203
I have a GE Humitrac 4-20mA output relative humidity, temperature sensor that I am using with the the NI 9203 board and the cDAQ 9174.  Currently I have the (+) from the power supply going to the V in (I1RH/P+) terminal of the sensor.  The RH RTN (I2RH/P-) terminal connected to AI0 of the 9203, the Temp RTN (I2-TMP) terminal is connected to the AI1 of the 9203, and the com (-) from the power supply connected to the com of the 9203.  Both channels are set up in MAX for 4-20mA.  I get zeroes on both channels when measuring ambient.  The manual for the sensor can be found here Humitrac Manual.
Any suggestions?  Thanks in advance.
LabVIEW 2016 - Windows 7
CLAD
(4,322 Views)
Hi there,
Just to make sure that you can read values other than zero, I would try to connect a known voltage value (like a battery) between your differential terminals. Also make sure that your differential terminals are not touching. You can use a Test Panel in the Measurement and Automation Explorer for simplicity.
Efrain G.
Visit http://www.ni.com/gettingstarted/ for step-by-step help in setting up your system.
(4,300 Views)
Thanks for the reply.  I'm not sure I understand your suggestion, the NI 9203 is a +-20 mA board.  Could you please elaborate?
Thanks.
CLAD
(4,287 Views)
Hi Clad,
Your connection looks OK to me. Have you selected the dip switch SW1 and SW3 on your sensor to current position? I would create a 4-20 current signal using a process meter and see if NI 9234 could see this signal if I were you.
(4,258 Views)
It was my scaling factors!  I created my curves using 4-20 instead of .004-.020, lesson learned Smiley Happy
CLAD
(4,256 Views)
That happened to me as well!
It's nice to hear you got your problem solved!
(4,252 Views)