Grounding unused/floating AI channels on PCI-6225
Hi, I'm currently running two DAQ cards on the same computer, PCI-6225 using 80 AI's and PCI-6723 using 32 AO's, that stimulate and record to/from a test system. This test system is a little unusual where sometimes both AI and AO channels can float and there may be no current passing to some AI channels. When reading single-ended AI's, if some AI’s float, I see a coupling between unused channels and used channels, i.e. if channels AI 1 to 4 are in use, and only AI 1 & 2 receive a signal (say 3V from AO), channels AI 3 & 4 take on the value of 1 &/or 2 (not 0V). This seems like a ground-loop problem, but after trying different ground configurations and source types using the “Signal source type” diagram in https://www.ni.com/en/shop/data-acquisition/measurement-fundamentals/field-wiring-and-noise-consider..., I’m just getting confused.
If I want the floating channels to go to ground should the unused channels have pull-down resistors to the AI's GND? Which configuration in the diagram should I use?
(3,041 Views)
What you are seeing is actually "ghosting" due to the MUX inside of the DAQ.  A resistor to GND would help, but you are technically draining a capacitor.  So you may have to play around a little to find a resistor value that works well for you.
(3,028 Views)
Thanks for the reply. After reading about ghosting here (https://knowledge.ni.com/KnowledgeArticleDetails?id=kA03q000000YHy6CAG&l=en-US), this makes some sense.
For the above example, I tried to lower the sample rate but the ghosting problem still persists -- maybe this is because of the high impedance/floating channels.
I'm going to look at using a Unity Gain Buffer (Voltage Follower) for each channel, as I would like to keep the sample rate high, and I can't really lower the input source impedance.
(3,014 Views)