Jay the Communist wrote:
>I am trying to configure a NI Lab-PC-1200 board, to collect analog
>input signal from a battery powered photodiode. Both multimeter and
>scope gave a DC reading of 1V. However when I attached it to the DAQ,
>it saturated at once and showed a straight line at 5V(high limit).
Hi Jay,
Is your card configured for single-ended, or differential measuring?
Single-ended measuring. provides you with 8 channels to be measured,
but forces you to use one single ground. It measures the voltage between
AGND and ACH0 up to ACH7.
Differential measuring allows you to measure 4 completely unrelated
signals, each with their own ground. You measure the signal between
ACH0 and ACH1, up to the signal between ACH6 and ACH7.
I
f you measure 4 signals or less, I'd recommend using differential
measuring. If I recall correctly, it can be configured in the Ni-Daq
control panel.
> signals remaining high for a long time
I don't have a hardware background either, but if differential measuring
doesn't solve the problem, if you put a big resistance between the two
channels, it will probably go down faster.
Hope it helps,
Walter van Iterson