High Speed Thermocouple measurement
I need to sample temperature at somewhere around 0.5-1kS/sec and am struggling to find a good solution.  The transitions I am monitoring occur on the scale of 0.1 sec and I need data density throughout that time to monitor the transition.
My thought is that a very fine gauge thermocouple will have the fastest response but am I'm finding that all of the ni daq solutions designed for thermocouples cap out at 100S/sec.  Has anyone else had this issue and what did you do to solve it?
(3,214 Views)
Hi AMillington - Why do you need to sample temperature at those speeds? Can you tell us more about your application?
Adena L.
Technical Support Engineer
(3,180 Views)
With TCs you deal with µV ... so you setup is prone to noise ... integration over a longer periode , usually at least one line periode 😉 , is the common easy way to fight noise.
So your setup needs to be calm.
More than one channel used ?  If Yes, avoid multiplexed cards!
An external CJ isn't the problem, so you need a DAQ with at least 1kSPS and some µV accuracy, or look for an external amplifier.
The 9251 (designed for sound 😉 fast, low noise , but high drift  ) ??
Resolution?
Accuracy?
Range?
Sorry, I tried to look up the the NI products, but the fast and good over view with a lot of  filters is gone, so best would be to contact a NI engineer ...
BTW: you can get really small and fast NTCs ..   It's just about self heating..
TC noise can be reduced if the TC is twisted!
I would use the 'my' HP3458 in direct samplemode with external CJ (hey, a devar with ice water isn't that bad) ..
Greetings from Germany
Henrik