01-25-2018 04:57 PM
I need to sample temperature at somewhere around 0.5-1kS/sec and am struggling to find a good solution. The transitions I am monitoring occur on the scale of 0.1 sec and I need data density throughout that time to monitor the transition.
My thought is that a very fine gauge thermocouple will have the fastest response but am I'm finding that all of the ni daq solutions designed for thermocouples cap out at 100S/sec. Has anyone else had this issue and what did you do to solve it?
01-26-2018 04:59 PM
Hi AMillington - Why do you need to sample temperature at those speeds? Can you tell us more about your application?
01-29-2018 06:50 AM
With TCs you deal with µV ... so you setup is prone to noise ... integration over a longer periode , usually at least one line periode 😉 , is the common easy way to fight noise.
So your setup needs to be calm.
More than one channel used ? If Yes, avoid multiplexed cards!
An external CJ isn't the problem, so you need a DAQ with at least 1kSPS and some µV accuracy, or look for an external amplifier.
The 9251 (designed for sound 😉 fast, low noise , but high drift ) ??
Resolution?
Accuracy?
Range?
Sorry, I tried to look up the the NI products, but the fast and good over view with a lot of filters is gone, so best would be to contact a NI engineer ...
BTW: you can get really small and fast NTCs .. It's just about self heating..
TC noise can be reduced if the TC is twisted!
I would use the 'my' HP3458 in direct samplemode with external CJ (hey, a devar with ice water isn't that bad) ..