02-22-2003 06:34 PM
02-25-2003
02:16 PM
- last edited on
09-11-2025
11:52 AM
by
Content Cleaner
Hello,
Even better than a timestamp is a hardware synchronization signal. When you perform an internally clocked acquisition with a DAQ board, the scan clock used is actually available as an output as the STARTSCAN which I believe is available on PFI 7. Consult your E-Series User Manual for more information. You can run a wire with this signal over to your other computer and use it as an external clock for it's acquisition. Then you are assured that both acquistions are synchronized. (Both will sample the data simultaneously.) There is also plenty more information on synchronization here in the Knowledge Base and Developer Zone i
f you're curious.
Russell
Applications Engineer
National Instruments