Multifunction DAQ
How to display voltage and current synchronously
Recently do a project about acquizeing voltage and current synchronouly ,then display. the design is welding a 0.5 ohm resistor(1%) in NI 6251 analog differential AI inputs, Then acquise the differential voltage at the AI input to convert to current. At the same time, I hope to grab the voltage of the analog input port. my issue is the current profile is no data. please help me .the program screenshot is show below. -------------------------------------------------------------------------------------------------- 1.may be you ask me why don't use no-dierential way to get voltage at all AI port, Then in software use minus to get the voltage at dierential AI inputs. this way will resolve my issue . i have a try ,but Current accuracy is fail. because my current expect range is 1uA~100mA. the samll current accuracy will not be promised if using single-way to acquise the voltage in NI 6251. NI 6251 Device spec(AI Absolute Accuracy) Nominal range(V) Absolute Accuracy(uV) Sensitivity 10 1920 112 5 1010 56 2 410 22.8 0.2 74 6.4 0.1 52 6.0 (i wanna use this range to measure samll current) ------------------------------------------------- Best regards Shawnh.Chen
Message 1 of 8
(5,341 Views)
Message 2 of 8
(5,333 Views)
1. You should just have 1 task for all of your analog inputs.
2. Use the DAQmx scale to do the math for you.
3. Reading 1uA will be very hard with a DAQ, even with some external circuitry.  I recommend the TSC103IPT.  It is specially made for amplifying current shunts.  The output is a single ended voltage.  It also has a configurable gain.  But even with the gain of 100, you are only going to have 50uV for your 1uA.  I doubt the DAQ can be that accurate.  Still with the gain of 100, your 100mA will give you 5V to your DAQ.
Message 3 of 8
(5,322 Views)
Thanks for your help.sorry for reply delay . get it, but I still have two confusion hope to get your reply 1.How to show the current waveform by accumulating 1 D Array data , Rather than show by every 100 acquisition points 2.What is the minimum identification unit of the amplifier TSC103IPT? I doubt that 0.5uA can be indentified?(0.5uV=0.5ohm*1uA) Because i didn't find anything about the auccary in TSC101IPT datasheet.
Message 4 of 8
(5,264 Views)
Accepted by shawn1
Use a waveform chart.  Charts keep a history, so you do not have to accumulate the data.  It does it for you in the display.
The amplifier is analog.  So it can gain up anything.  But you really should think about using a larger resistor if you really want to measure that small of a current.
Message 5 of 8
(5,254 Views)
Thanks Crossrulz. The measured current will be not accurate if i use a larger resistor (I have a try using a 1 ohm resistor to get the measured current 5uA ,but i try a 0.1 ohm resistor to get measured current 2uA. other example like this ). So i only just use a smaller resistor to reduce loading influence. Another question, You just kindly remind me to use a amplifier TSC103IPT.and said the amplifier is analog to gain up anything,but i doubt that The NI 6251 analog inputs is also analog ,why it can't gain up angthing? Haha.My question is so much .
Message 6 of 8
(5,230 Views)
shawn1 wrote:  but i doubt that The NI 6251 analog inputs is also analog ,why it can't gain up angthing?
The 6251 uses an Analog To Digital Converted.  That is what has a minimum sensing voltage.  That is also where the resolution specification comes from.
Message 7 of 8
(5,223 Views)
Get it ,Thank you very much .
Message 8 of 8
(5,203 Views)
