Hello Kostya,
Connecting 30-40 volts to your PCI-6025E would not be recommended. The PCI-6025E has a maximum working voltage of 11V according to the specifications. This means that the common mode voltage+signal should not exceed 11 V on any terminal. For this measurement you will need a card with a maximum working voltage which meets the requirement of your signal or external connections to bring the signal within the working voltage of the card.
Two possible solutions include using another DAQ board which has isolation such as the PCI-6230 or using signal conditioning such as an SC-2345 and SCC-AI01.
Please refer to High-Voltage Measurements and Isolation - General Analog Concepts for information about isolation and working voltages.
Hopefully this helps,
Jennifer O.
Applications Engineer
National Instruments