How to set input range to 100mv

I use AI_configure(....) to set input range.And i want to set input range to 100 mv .How should i do.
Please tell me..........my DAQ is PCI MIO 16E-4
0 Kudos
Message 1 of 2
(2,642 Views)
You will need to set the gain of each channel using the AI_Read function. The input range parameter of AI-Configure is acutually ignored for E Series devices. The polarity parameter will allow you to set the range to unipolar (0 to 10 V) or bipolar (-5 to +5).

You use the gain parameter of the AI_Read to futher specify the input range. The valid gain settings for the 16E-4 can be found on page A-2 of the appendix in the user manual linked below. For a +/- 100 mV range you would set the polarity of AI_Configure to bipolar and the gain parameter to 50 of the AI_Read. I hope this helps.