How to use a thermistor as a CJC for thermocouple measurement ?
I am using NI-USB-6211, for temperature measurement via Thermocouples (Type J), and of course i have to introduce the CJC value to have the accurate temperature values.
Ni-USB-6211 does not have a built-in cold junction compensation. and i have to take measurements continuously through a year.
How can i use a thermistor as a Cold Junction Compensation (CJC through a channel)  for my measurements with the thermocouples ??
thank you.
(3,073 Views)
To measure a thermistor, you need an accurate current source and something to digitize the resulting voltage drop across the resistive element.  The 6211 cannot do the former.  It could do the latter, with its accuracy depending on the accuracy of the 6211's AI (not great).  You'd also have to control the thermistor's position near the cold junction compensator and, ideally, isothermally isolate it, which would involve building some kind of enclosure.  Even with all of that, you're going to have several degrees of measurement inaccuracy before also including thermocouple and thermistor measurement inaccuracies.
It would be way easier to just buy a DAQ device that is designed to measure thermocouples with built-in CJC compensation.  These devices will have an accuracy statement for the entire measurement solution.  You just have to add the error of your thermocouple.  If you're looking for an NI solution, the TC-01 is a single channel, cheap device.  You could also use a single slot cDAQ like the 9171 with a thermocouple module (9211, 9212, 9213, etc.) if you need multiple channels.
(3,005 Views)