I need a data acquisition system that can supply an output of 2.5-5 volts.
I have a sensor that requires a low current (2-7mA) 2.5-5 volt excitation voltage for 5ms, and then it outputs a signal that is approximately 10-40% of the input voltage. I need a National Instruments data acquisition system that is capable of providing the excitation voltage and then measuring the output.  I currently have a SCXI 1000 Chassis that is connected to a PC with a PCI M10 16XE 10 and a TBX 1303 terminal block. If possible I would like to just get a new module for the Chassis and a new terminal block. Any suggestions?
(3,131 Views)
Your existing DAQ card can source the 2.5 to 5 volts Analog output, on any of its 2 AO channels
You speak of TBX 1303 in your system.
Do you have a SCXI module in your SCXI chassis, probably SCXI 1100 or SCXI 1102 which uses this terminal block??
(3,119 Views)
Yes, it is a SCXI 1102 module. I left that part out in my system description.
(3,111 Views)
You could use one of the free Analog input channels in the DAQcard itself, to measure the response as well.
If not, you could use a SCXI channel ( with range =/- 10V , or gain set to 1) for measuring this response voltage
(3,075 Views)