I'm using a USB-6009 in conjunction with a K-type thermocouple and AD597 conditioner/amp. I initially had 4 signals plugged in which correctly reported the 10mV/degC signal using the differential setting, however, I needed to add one more reading to the experiment. Since the USB-6009 only has 4 differential channel and 8 RSE channels, decided to make the switch to accommodate the 5 readings. I changed all of the inputs to the correct terminals (The negative output lead of the amps are all common and connected to the USB-6009 ground and the positive leads are connected to the AI# terminals) and switched the Labview settings to RSE, but all the readings in Labview went to ~0.33V. This result was indifferent to whether the amps were powered by battery, outlet or even unplugged as well as if the thermocouples themselves were disconnected. A multimeter across the terminal port shows the same ~0.33V. If the leads of the powered amps are disconnected from the USB-6009 terminals, the multimeter shows that the signals from the thermocouple amp are correct again (room temp -> ~0.22V) and adjust to heat changes. Returning everything back to differential mode also corrects everything again. We have several USB-6009 units so I disconnected everything and retried on a different USB-6009 and obtained the same erroneous results.
06-23-2010
11:03 AM
- last edited on
07-21-2025
03:04 PM
by
Content Cleaner
Hi jasonjps,
When using the USB-6009, it is not uncommon to get noisy results, as mentioned in this knowledgebase article. I would recommend moving to another form factor like the USB-9211. With an RSE configuration, you will be even more subjected to noise than a differential configuration.
06-23-2010 12:03 PM
It seems that this is not a noise problem. The RSE voltage is virtually static at ~0.33V. It's not like it's near the correct value and reacts to temperature changes with a large margin of error.
06-24-2010
04:12 PM
- last edited on
07-21-2025
03:04 PM
by
Content Cleaner
Hi jasonjps,
Actually, you should generally not use an RSE configuration for such a low voltage input value, as noted in the following knowledge base:
Field Wiring and Noise Considerations for Analog Signals.
There is a warning note in the knowledge base that states:
Single-ended input connections can be used when all input signals meet the following criteria.
Differential connections should be used when any of the above criteria are violated.
10-10-2011 10:18 AM
Hello All,
I am still confused on RSE and Differential Mode. I have a pressure sensor with a measurement range of +/-5V which I want to connect to USB-6009. Since there is only one pressure transducer I attached the positive end to AI1 and the negative end to ground. Is this RSE or Differential mode? (If I select Differential Mode in MAX, the signal is less noisy then in RSE).
Thank you
greentea2
10-11-2011
11:18 AM
- last edited on
07-21-2025
03:05 PM
by
Content Cleaner
Hello,
This is wired in an RSE configuration, as you have the positive input wired to the AI0 and the negative input wired to ground. The field wiring guide (https://www.ni.com/en/shop/data-acquisition/measurement-fundamentals/field-wiring-and-noise-consider...) does a good job of explaining how the different types of measurements are wired. A differential measurement would require wiring the positive terminal to AI0+ and the negative terminal to AI0-.
Best,
Dan Nelson
Applications Engineer