Interfacing a differential probe to NI USB-6281
We have a differential probe designed to be connected to an oscilloscope, so the connection to the scope is 1 MOhm.
We want to connect this probe to a NI USB-6281, what is the correct way to do it ?
Please post your comments/suggestions.
David.
(4,528 Views)
Hi DavidDLC
Did you try NI MAX?
When you create a task with the device you are using,  you can use the connection diagram on NI MAX in order to see the connection.
You can check this document:
https://knowledge.ni.com/KnowledgeArticleDetails?id=kA03q000000x4q4CAA&l=en-US
Warm regards,
Sigi D.
Application Engineer
"everything is achieveable through technology'
(4,500 Views)
Thank you for your very good recommendation.
The way we did it was to put 1 MOhm resistors in series with the differential probe, it worked very well, we compared the measurements against an scope and the results are spot on.