05-28-2015 06:11 PM
We have a differential probe designed to be connected to an oscilloscope, so the connection to the scope is 1 MOhm.
We want to connect this probe to a NI USB-6281, what is the correct way to do it ?
Please post your comments/suggestions.
David.
05-29-2015
03:04 PM
- last edited on
10-20-2025
12:20 PM
by
Content Cleaner
Hi DavidDLC
Did you try NI MAX?
When you create a task with the device you are using, you can use the connection diagram on NI MAX in order to see the connection.
You can check this document:
https://knowledge.ni.com/KnowledgeArticleDetails?id=kA03q000000x4q4CAA&l=en-US
Warm regards,
Sigi D.
Application Engineer
05-30-2015 01:19 AM
Thank you for your very good recommendation.
The way we did it was to put 1 MOhm resistors in series with the differential probe, it worked very well, we compared the measurements against an scope and the results are spot on.