I am using
the a new NI USB 6008 DAQ. When connecting the Ao0 & GND with Ai0 + & -
respectively and measure in the test-panel whether the before set voltage for
the output appear at the input the result differ systematically from what I
would expect. 
- the device passed the self-test
- I am using the analogue-side 
- channels are set correctly
- output is set to DC, 0 min & 5 max, the voltage itself is set to a value
between 0 and 5
- input is set to 0 to 5 volt again, differential measuring, single mode
The whole procedure is described also here:
http://techteach.no/tekdok/usb6009/#test
Result:
For every 0,5 V the output rises the input increases by ~0,14 V
This is what I get, on the left the generated output (V), on the right the
displayed input (V).
0     -     1 
0,5   -   1,14 
1     -    1,28 
1,5   -   1,43 
2   -   1,57 
2,5    -   1,71 
3   -    1,85 
3,5   -   1,99 
4   -   2,14 
4,5   -   2,28 
5   - 2,42 
No scaling was defined whatsoever. When measuring the generated output with an
external device the correct numbers appear.
Any help or explanation would be greatly appreciated as I am just starting with
these matters.