MY NI 9219 module rails at 60 V

How do I troubleshoot(fix/repair) my NI 9219 module that rails at +60 V when its channels are left open(-circuit) ?
0 Kudos
Message 1 of 5
(3,513 Views)

ft-lbs: 

 

Perfectly normal, the front end input circuitry is building up a charge (it's so low in capacity it won't hurt you).

Short the inputs together, it should read 0. Connect a source to it, it should read the correct voltage.

 

-AK2DM

~~~~~~~~~~~~~~~~~~~~~~~~~~
"It’s the questions that drive us.”
~~~~~~~~~~~~~~~~~~~~~~~~~~
0 Kudos
Message 2 of 5
(3,509 Views)

AK2DM

 

I've actually done those (shorted the inputs together and connected to a voltage source), but the device is unresponsive. And this is true for all 4 channels.

 

-Torque in ft-lbs

0 Kudos
Message 3 of 5
(3,505 Views)

N-m:

 

That does not sound normal then. Hopefully someone from NI will chime in, they will probably determine that you will have to send the module in for repair.

Have you tried cycling power, and was it previously working?

 

-AK2DM

~~~~~~~~~~~~~~~~~~~~~~~~~~
"It’s the questions that drive us.”
~~~~~~~~~~~~~~~~~~~~~~~~~~
0 Kudos
Message 4 of 5
(3,498 Views)

torque,

 

It does sounds like the device is damaged, how have you been getting these Voltage measurments?  Are you using a LabVIEW program or test-panels to check these values. If you need to get the device repaired, you will need to call our Customer Service number 800-531-5066, and we will be able to give you details on getting your device repaired. Thanks!

Aaron W.
National Instruments
CLA, CTA and CPI