Multifunction DAQ

cancel
Showing results for 
Search instead for 
Did you mean: 

Maximum Input current for 6210 /-10V Range

Solved!
Go to solution

What is the input current spec for the NI USB 6210 when making a single ended (i.e one of its 16 channels) reading in the +/-10V range? I need to quantify the error in a voltage divider I am using to measure a 200VAC ground referenced signal, since part of the current through the voltage divider is splitting off to drive the ADC input on the 6210.

0 Kudos
Message 1 of 2
(2,871 Views)
Solution
Accepted by topic author @dave@

From the specs document:

 

Input impedance

Device on

AI+ to AI GND ......................>10 G

with 100 pF

AI– to AI GND.......................>10 G

with 100 pF

Device off

AI+ to AI GND..................1200

Ω in parallelΩ in parallelΩ

AI– to AI GND ..................1200

Ω

Input bias current.............................±100 pA

 

I presume the +/-100pA is from 10V/10Gohms. I doubt that the 6210 will load down the divider.

 

-AK2DM

~~~~~~~~~~~~~~~~~~~~~~~~~~
"It’s the questions that drive us.”
~~~~~~~~~~~~~~~~~~~~~~~~~~
0 Kudos
Message 2 of 2
(2,869 Views)