Hi Alexandr,
Good afternoon and I hope your well today.
Thanks for the post.
You are correct about the NI 9215 - you shouldn't see a delay between channels readings - so are you reading the same signal into both channels, also try to get the wires the same length, as it will take a finite amount of time for the signal to travel from the sensor to the 9215. ( NI 9215 has a group delay of 0 samples(not a delta sigma ADC)).
If its not the same signal, how do we know the phase delay isn't in the two signals? For example Connecting a Floating Differential Voltage Signal to the
NI 9215 with Screw Terminal requires a bias resistor.
User Manual
http://www.ni.com/pdf/manuals/373779e.pdf
You can account for the delay - if you haven't got the time to find the cause of the delay.
Something to keep in mind is that the external circuitry can introduce additional skew. In what sort of region is your phase delay? - I'd have to check it up but I think you'd be looking at ns for interchannel delay.
Are you using code to generate the external signals? Maybe the labVIEW code is generating a phase delay..
Please let me know your thoughts, and I'd be more than happy to discuss this issue with you.
Hope this finds you well,
I would also make sure you've input circuitary is correct as per the manual,
03-04-2009 08:29 AM
03-04-2009 08:40 AM
Hi Alexandr,
Thanks for the reply.
Your setup for the C Series module looks correct - which is good.
I would suggest cable length being the main cause - you can add a filter to increase a delay in the shorter cable, here is a knowledgebase article which mentions this... it is more with adding a group delay for modules with different specifications, but same sort of thing.
How to Compensate for Different Group Delays with C Series Modules in LabVIEW FPGA
http://digital.ni.com/public.nsf/allkb/74EB238E1BCADD528625735300681A7D?OpenDocument
Otherwise, I would maybe consider altering the phase in software.
Note: The delay of around 1ns from what I have read is pretty good - but there is no group delay with the module, so it must be your setup I am afraid.
Sorry I didn't have any other thoughts from the cable length.
Hope this helps,
03-04-2009 10:22 AM
The different channels in the 9215 will have slightly different delays, since they have slightly different bandwidths. I would expect variation in the 10's of nanoseconds, but it could be more. However, 200m of cable would introduce delay on the order of 1 usec, so I suspect that's your primary culprit.
You could try connecting the exact same signal to your two channels and measure the phase delay between them to see how much of the error is due to the module.
Chris
03-04-2009 10:55 AM
Hi Alexandr,
Please let us know how you get on,
Chris is correct - with the idea of using the exact signal for two channels, I was trying to suggest this earlier on..
I hope this helps resolve your issues with the delay,