NI 9219 to measure temperature using thermistor in 4-wire mode
I am measuring temperature with thermistor sensor using NI 9219 AI module on a cDAQ 9178 chassis. There is no issue when measurement is done using 2-wire mode and displayed result is in complete agreement with calibration report provided by Thermistor manufacturer. However we want to take measurement over a long distance. To get accurate result, we used 4-wire resistance measurement scheme. But the displayed result shows Zero resitance (short). The connections are proper and checked multiple times. I am not able to figure out the error. Whether NI 9219 cannot measure Thermistor resistance in 4-wire mode.
At 25 deg C, Rthermistor = 10k.
Need help.
(4,685 Views)
Accepted by arihant
Hi, what do you mean by "long distance"? What type of wires do you use?
Does the connection look like this?
Also - the program you sent is configured to use 2-wire, not 4-wire method.
(4,653 Views)
My bad.
I switched the Pin 4 connection with Pin 6. I will try after rectifying it.
Thanks for your nice connection visual.
(4,635 Views)
@arihant
were you able to get this working? What is the accuracy you are getting?
Bhavin
(3,809 Views)
Sorry as I could not reply.
Yes, I was able yo do it.
(3,328 Views)
Please mark the appropriate reply as Solution if already solved.
Santhosh
Soliton Technologies
(3,310 Views)