Multifunction DAQ
NI 9269 Offset by 13.68 V
So I'm using a NI 9269 for generating DC voltage. For the a01 and a03 channels, my outputs are offset by 13.68 V. Is this some kind of calibration issue? If so, how do I rectify it?
A R
Message 1 of 12
(5,698 Views)
Is it truely an offset or is that just the maximum it can go to?  Like if you told it 5V it goes to 18.68V.
My first thought is that this is a grounding/wiring issue.  After that, I would say there is something wrong with your module and you should call up your local NI rep and ask them what can be done for repair and/or swapping out modules.
Message 2 of 12
(5,683 Views)
What's it powered off of?  Sounds strangely close to automotive charging voltage.
For starters I'd say check the connections and grounds.
Message 3 of 12
(5,674 Views)
Yup it's a true offset. If i set it to 5V it goes to 18.68, set to 10 I get 23.68 (although the limit of the device is +/- 10V).
Message 4 of 12
(5,670 Views)
It's powered off of USB through the cDAQ chassis.
Message 5 of 12
(5,669 Views)
Since the outputs are stackable to get up to 40 volts, you really need to provide the exact wiring that you have done.
Message 6 of 12
(5,662 Views)
Then you have a wiring issue.  Do you have a diagram of exactly how you have it wired up?
Message 7 of 12
(5,645 Views)
The wiring isn't too complicated. The module is used to generate DC voltage. There is + and - terminals. I have these hooked up directly to the DMM.
Message 8 of 12
(5,639 Views)
If it's not that complicated, why can't you attach the diagram? You said you were using multiple modules and its important to know if you've somehow connected one to another. This is a simple request.
Message 9 of 12
(5,633 Views)
I didn't attach a diagram cuz I didn't see how it would help. Essentially, the four channels are used as four different voltage sources. They are not connected together.
Message 10 of 12
(5,629 Views)