06-16-2008 01:11 PM
Solved! Go to Solution.
06-17-2008 07:08 PM
06-17-2008 11:22 PM
06-18-2008 03:56 PM
Hi jinwei,
Yes, you should keep your SCB-68 in Differential Switch Configuration as indicated in Figure 4-2 of your user manual.
S_Hong
06-18-2008 07:31 PM
06-19-2008 10:07 AM
06-19-2008 02:24 PM
06-20-2008 11:52 AM
HI jinwei,
I'm sorry for the confusion. It appears that the PCI-6221 does not generate output currents and I failed to notice this. It seems that your method was the way to go on this one. If you have any questions on how the system can be connected, you can always use the Connection Diagram tab in you Resistance task in MAX to help you.
06-20-2008 12:21 PM
06-23-2008 05:43 PM
Hi jinwei,
It seems to me that this is expected behavior as you are trying to measure very small voltages. From my calculations, if you are using a current of 150 µA and putting it through a 25 ohm resistor, you should be expecting to measure 3.75 mV. The accuracy of your measurement will depend on many factors such as when the device was last calibrated, temperature, the number of readings you are taking, etc... By using the Accuracy Calculator on our website, you can have a good idea of how accurate your readings will be. I calculated that at best, even with a 100 point average, you will achieve an absolute accuracy of approximately +/- 0.0709 mV which translates to approximately 0.472 ohms.