Multifunction DAQ
NI USB-6251 anti-aliasing filters
Hi
I have NI USB-6251 and I have to use maximum sampling rate 1250 Khz how i could put anti-aliasing filters? help please
also I have NI 9234 to receive a signal from the 4-Channel with sampling rate 50 khz for each channel should I put anti-aliasing filters in this case or not?
thanks
(3,932 Views)
Accepted by topic author EngYasir
The USB-6251 can only sample at that rate on one channel.  If you use multiple channels the maximum aggregate rate is 1 MS/s. You should install an anti-aliasing filter before the input to the device.  The filter bandwidth could be as low as 625 kHz ( the Nyquist frequency) or somewhat higher depending on the noise or interference which accompanies your signal.
The NI 9234 has a built in anti-aliasing filter. Since your samplerate is just slightly below the maximum sampling rate, you should not need any other filter. If you have an extreme case of a strong interfering signal just outside the signal bandwidth, a special filter might be useful
Lynn
(3,924 Views)
thank v.
(3,885 Views)