10-07-2013 06:33 AM
Hi
I have NI USB-6251 and I have to use maximum sampling rate 1250 Khz how i could put anti-aliasing filters? help please
also I have NI 9234 to receive a signal from the 4-Channel with sampling rate 50 khz for each channel should I put anti-aliasing filters in this case or not?
thanks
Solved! Go to Solution.
10-07-2013 02:31 PM
The USB-6251 can only sample at that rate on one channel. If you use multiple channels the maximum aggregate rate is 1 MS/s. You should install an anti-aliasing filter before the input to the device. The filter bandwidth could be as low as 625 kHz ( the Nyquist frequency) or somewhat higher depending on the noise or interference which accompanies your signal.
The NI 9234 has a built in anti-aliasing filter. Since your samplerate is just slightly below the maximum sampling rate, you should not need any other filter. If you have an extreme case of a strong interfering signal just outside the signal bandwidth, a special filter might be useful
Lynn
10-09-2013 09:17 PM
thank v. johnsold