08-31-2011 07:55 AM
Hello All,
I'm using a PCI-6031E for an application with multiple analog inputs that are typically in the +/-10VDC range. On one particular channel, when I apply approx -9VDC to the channel, a lot of "noise" will appear (what seems like as much as +/-0.5V). The noise is visible within my LabVIEW application, and can be verified with a scope. The interesting thing is that the noise disappears if I stop scanning all DAQ channels and read only the channel of interest. Likewise, if I move the signal to a different channel the noise disappears, even when scanning. Finally, I can typically make the noise disappear by merely disconnecting and reconnecting the source. Any thoughts as to what to explore? Has anyone encountered anything similar?
09-01-2011 04:54 PM
Hi Greg,
Could you do a FFT and see what frequency the noise is at.
Josh Brown
		09-01-2011
		05:54 PM
 - last edited on
		07-20-2025
		05:16 PM
 by
			Content Cleaner
Hi Greg,
I have some more questions/suggestions. Do you have your channel connected single-ended or differentially. Refer to table 1 in this knowledge based article if you need to.
Also, have you tried connecting a battery to the bad channel? If not could you do so and report back what you see. If the battery gives you a bad signal then this may suggest a bad channel.
Josh Brown
09-02-2011 06:24 AM
Sorry that I forgot these important details...
-FFT shows the frequency to be in the MHz range
-Configuration is NRSE
I have two seperate identical systems that both exhibit the same issue on the same channel- thus I think a bad channel is not likely.
Thanks!
Greg
09-06-2011 10:50 AM
Hi Greg,
Could you try configuring the system in the differential format and see if the noise still exist?
Josh Brown