Noisy Channel on PCI-6031E
Hello All,
I'm using a PCI-6031E for an application with multiple analog inputs that are typically in the +/-10VDC range.  On one particular channel, when I apply approx -9VDC to the channel, a lot of "noise" will appear (what seems like as much as +/-0.5V).  The noise is visible within my LabVIEW application, and can be verified with a scope.  The interesting thing is that the noise disappears if I stop scanning all DAQ channels and read only the channel of interest.  Likewise, if I move the signal to a different channel the noise disappears, even when scanning.  Finally, I can typically make the noise disappear by merely disconnecting and reconnecting the source.  Any thoughts as to what to explore?  Has anyone encountered anything similar?
(3,169 Views)
Hi Greg,
Could you do a FFT and see what frequency the noise is at.
Josh Brown
(3,148 Views)
Hi Greg,
I have some more questions/suggestions. Do you have your channel connected single-ended or differentially. Refer to table 1 in this knowledge based article if you need to.
https://www.ni.com/en/shop/data-acquisition/measurement-fundamentals/field-wiring-and-noise-consider...
Also, have you tried connecting a battery to the bad channel? If not could you do so and report back what you see. If the battery gives you a bad signal then this may suggest a bad channel.
Josh Brown
(3,145 Views)
Sorry that I forgot these important details...
-FFT shows the frequency to be in the MHz range
-Configuration is NRSE
I have two seperate identical systems that both exhibit the same issue on the same channel- thus I think a bad channel is not likely.
Thanks!
Greg
(3,140 Views)
Hi Greg,
Could you try configuring the system in the differential format and see if the noise still exist?
Josh Brown
(3,104 Views)