There are 2 physical connectors but all analogs go back to a single A/D converter and share the same grounds(AI Sense for NRSE or AI GND for RSE). 15V is right at the limit for this device's overvoltage, could have broken something.
Look at Table 1 in the Field Wiring Guide. Which configuration are you trying to use, and do you have the correct biasing resistors? Can you probe on the SCC with your real signal attached and see how it looks?
09-24-2012 02:13 PM
09-24-2012 02:43 PM
The AI on a DAQ card should not be pulling a power supply up to 10V, something's definitely wrong there. Try ohming out the path and make sure nothing's shorted on the SCC. Is your power supply isolated? Not just that it says isolated, but actually is? Nobody added a ground/chassis wire or an internal short to case? Are you using +/-5V scale? Could be railed low if so, or the common mode voltage high enough to push the signal out of range.
You could try connecting signal with the DAQ unpowered, if you still read -5V with a meter on the unpowered DAQ setup you know there's a hardware problem somewhere. As long as you keep it within +/-10V input you won't damage the card.
09-25-2012 09:34 AM
04-01-2013 07:05 AM
Iam using NI pci 6259 M series DAQ card with bnc 2110 connector. My problen is that iam unable to read analog input in measurement and auotmation explorer.I have installed driber softaere and hardwre but with no success. The reading shown in measurement and auotmation explorer is 3.11 volt in all analog channels irespective of analog input. Any help will be greatly appreciated. Y Tyagi
04-02-2013 09:14 AM
Tyagiy,
I would recommend creating a new forum post for your question so that this thread does not get off topic and so that you get more responses to your question.