Dear all,
I would like to meassure some Current inputs 4-20 mA and some PT1000 Sensors. Is this possible with the PCIe 6323 card? There is no information about that in the handbook.
Thank you in advance!
Michael
(2,164 Views)
Accepted by topic author MichaelG2
RTD requires an excitation current to work properly. You should get a proper RTD measurement device. Measuring Temperature with Thermocouples, RTDs, and Thermistors
https://tmesystems.net/
(2,154 Views)
@MichaelG2 wrote:
Dear all,
Michael
6323 is Analog Voltage input card and not best suited for the 4-20mA measurement or PT1000. You could measure it with additional external circuit but it could be extra work, instead but the right hardware for the job.
Santhosh
Soliton Technologies
(2,106 Views)
Is there a PCIe card that can handle a current measurement signal?
Michael
(2,053 Views)
Michael
Unfortunately, no PCIe equivalent. Here are the available models - https://www.ni.com/en-us/shop/category/current.html?page=1&foc=0
Instead, you can also use an external current-to-voltage converter.
Santhosh
(2,049 Views)