PXI-6289 DAC takes long time to set
I am using a PXI-6289 and a PXIe-8101 controller. The controller is setup with a real-time OS. Recently I moved from LabWindows/CVI 2012 to 2013. When I made changes to my application on the PXI I had to update the software on the PXI system so it would be compatible with the software in the LabWindows/CVI. After I did this I recompiled my PXI application and noticed some undesirable behavior. I finally narrowed it down to the part where I am setting the DAC on the PXI-6289. I noticed it is taking 80ms to change the DAC output. I am thinking I might have some software installed which has some compatibility issues with the realtime OS on the PXI. The software installed is the NI-DAQmx 9.7.5. Has anyone had similar problems?
It probably would not be noticeable in most applications but I can see it because I am measuring the transient response of an RC filter so I am trying to catch the DAC setting change while I am acquiring a waveform with the DUT. Like I said this only started happening after I moved from LabWindows/CVI 2012 to 2013 so I think it might be the software. Below is a snippet fo the code which sets the DAQ; maybe there is a better way of doing this.
DAC_statusMessages_t SetDACoutput(double output)
 DAC_statusMessages_t status = DAC_OUT_STATUS_OK;
 float64 VoltageLevel;
 char ErrorMessage[DAC_ETHERNET_MESSAGE_LEN];
 char DAQ_ErrorMessage[DAC_MESSAGE_LEN];
    TaskHandle taskDACOut;
 int32 returnDAQmx_status;
 char DeviceName[AO_6289_CHAN_TEXT_SIZE];
 strncpy(DeviceName, PXI_6289_NAME_STR, sizeof(DeviceName));
 strncat(DeviceName, DAC_OUT_CHAN_0, sizeof(DeviceName));
 VoltageLevel = (float64) (output);
 returnDAQmx_status = DAQmxCreateTask("DACOutTask",&taskDACOut);
 if( 0 != returnDAQmx_status)
  strncpy(ErrorMessage,"DAC out Error: ", DAC_ETHERNET_MESSAGE_LEN);
  DAQmxGetErrorString (returnDAQmx_status, DAQ_ErrorMessage, DAC_MESSAGE_LEN);
  strncat(ErrorMessage,DAQ_ErrorMessage,DAC_ETHERNET_MESSAGE_LEN);
  strncat(ErrorMessage,"\n",DAC_ETHERNET_MESSAGE_LEN);
  SendRT_ErrorMessage(ErrorMessage);
  // Clear the task
  DAQmxClearTask (taskDACOut);
  return DAC_OUT_STATUS_NOK;
    returnDAQmx_status = DAQmxCreateAOVoltageChan(taskDACOut,DeviceName,"",-10.0,10.0,DAQmx_Val_Volts,"");
  // Clear the task
    // DAQmx Start Code
    returnDAQmx_status = DAQmxStartTask(taskDACOut);
  // Clear the task
        // DAQmx Write Code
    returnDAQmx_status = DAQmxWriteAnalogF64(taskDACOut,1,1,10.0,DAQmx_Val_GroupByChannel,&VoltageLevel,NULL,NULL);
  // Clear the task
 g_SetDACActual = GetTimeUS();  // get the time to sort out how long this function is taking to finish
 // Clear the task
 return status;
(4,357 Views)
I would recommend downloading and installing DAQmx 9.9 and then installing DAQmx 9.9 to the PXI controller to see if that changes the output.
Let me know if this helps!
Rob B
FlexRIO Product Manager
(4,335 Views)
Accepted by topic author DPearce
I found that the SetDACoutput() function was taking 80ms to complete. I fixed it by splitting up this function into three functions
Create and setup the task
Set the DAC
Clear the task
I create and setup the task at the beginning of my application. And clear the task at the end of my application. During the test I only cal DAQmxWriteAnalogF64().
This worked out well. Still do not understand why the time changed so dramatically in LabWindows/CVI but the work around is better anyway.
(4,329 Views)