Multifunction DAQ
PXIe 6341 and SCB-68: problem with multiple inputs.... *URGENT*
I have an urgent problem that I'd really appreciate some help with.
I have a terminal block (SCB-68) connected to a multifunctional DAQ (6341). There are 4 inputs in the terminal block:
1. Laser displacement sensor, coming from a controller as an analog-out voltage signal ("1V" and "0V" on the controller). Connected to AI 0 and AI GND.
2. Accelerometer (ICP, coming from a signal conditioner so as voltage) connected to AI 10 and AI GND.
3. Load cells (ICP, coming from a signal conditioner so as voltage) connected to AI 11, 12, 13 and 14, each connected to a AI GND pin nearby.
Problem:
Via MAX, if I read in the above 6 channels from my DAQ (6341), my displacement sensor (#1) and accelerometer (#2) behave fine. However, two of the load cell channels (AI 11 and AI 12) shift in voltage, as if they're following the displacement sensor voltage (#1). For example, there's typically a -800 mV shift in AI 11 voltage.
This problem totally disappears when I don't request the laser displacement (AI 0, #1) reading. i.e. as soon as I remove that channel from my task, the load cell and accelerometer readings go back to normal (base voltage ~0V). Again if I re-request the laser channel, I see a vertical shift in 2 of the load cell readings (the shift varies with the current voltage in AI 0).
Attempts at fixing:
-tried changing the laser sensor to differential by using AI 0 and AI 9
-tried removing the ground from the laser 120V power supply
Any ideas at all would be greatly appreciated!
SCB-68:
https://www.ni.com/en-us/support/model.scb-68.html
PXIe-6341:
https://www.ni.com/en-us/shop/model/pxie-6341.html
(4,416 Views)
I'm not entirely sure, but to me it sounds like the ICP channel has high impedance which is causing the channel "ghosting". I would recommend taking a look at this document for suggestions of some things you might try:
https://knowledge.ni.com/KnowledgeArticleDetails?id=kA03q000000YHy6CAG&l=en-US
(4,388 Views)
Interesting Neil! I have a feeling you're right.
I plan to read the article tonight, but I'd like to ask:
Do you think the problem will be solved if I use more than 1 signal conditioner for my 4 cells? e.g. 2 cells in 1 signal conditioner and 2 cells in another. Not sure why/how that would solve the problem, but that looks like 1 of the things that would be simple to implement quickly.
(4,384 Views)
The article Neil linked to states that changing the order of the channels to avoid ghosting only works if I am "not sampling in round-robin mode." What does this mean?
If I understand correctly what's happening is my 1st low-impedance displacement sensor is "ghosted" in the high-impedance 2nd and 3rd load cell readings. To avoid this, can I put an empty short-circuited channel between the displacement sensor and the load cells?
displacement sensor: AI0/GRND
empty short-circuited: AI1/GRND <<<<< absorbs the discharge from multiplexer charge
load cells: AI9/GRND
etc.
or
load cells: AI0/GRND
displacement sensor: AI9/GRND
But I'm not sure if this helps anything...again the article (linked again below) says this only helps if I'm not sampling in round-robin mode.
Thanks for the help!
(4,378 Views)
Accepted by ap8888
That is the general idea.  If you must use sources with high impedance and you have enough channels available, placing a grounded-input channel between each of the others usually helps.  You must scan those grounded channels.  Just grounding them is not enough.
Example: Sensor A = AI0, Ground = AI1, Sensor B = AI2, Ground = AI3, Sensor C = AI4, Ground = AI5.  Read AI0:5.
Lynn
(4,362 Views)
Makes sense Lynn.
And "ground" is equivalent to shorting the analog input with a ground.
So I can connect the following and read in data in the following order:
Sensor A = AI0/GRND
Shorted = AI1/GRND
Sensor B (high impedance) = AI10/GRND
Sensor C (high impedance) = AI11/GRND
Sensor D....
Sensor E....
Sensor F....
Shorted = AI7/GRND
Seem like a remedy?
(4,355 Views)
I would put a grounded input between each of the high impedance sensors.
AI0 = Sensor A
AI1 = Ground
AI2 = sensor B
AI3 = Ground
AI4 = Sensor C
AI5 = Ground
AI6 = Sensor D
AI7 = Ground
AI8 = Sensor E
AI9 = Ground
AI10 = Sensor F
AI11 = Ground
Common or return side of all sensors connected to ground.
Set up the read task to read AI0:11. Then discard the odd numbered channels (the grounded inputs).
Lynn
(4,351 Views)
