Hi Andreas,
If you are connecting two poles, i.e. + and 0, then you should connect the 0-signal to channel AI_x+8 and select the differential terminal configuration. If you connect the +-signal to AI_3, the 0-signal must be connected to AI_11. If you have chosen to measure RSE instead, you have connected the signals correctly. But RSE is big no-no if the 0-signal is grounded. To get rid of ghost signals on channels not connected, you should connect all to AIGND. For more information please have a look at the following link. Data Acquisition: Troubleshooting Bad Readings
Hope it helps