Hi,
I think that I can shed some light on what may be happening with your thermocouple readings. From the SCC-68 User Guide I have found the following information.
Temperature differences between the cold-junction temperature sensor and
the actual cold junction formed by a thermocouple connected to the screw
terminals will directly contribute to thermocouple measurement errors.
With minimal loading on the SCC-68 ±15V power supply, the temperature
difference between the screw terminals of the SCC-68 is typically less than
1 °C. Variations in the ambient temperature, SCC module heat dissipation,
and heat conducted along thermocouple wire can all contribute to
isothermal errors across the screw terminals.
For more information on thermocouple theory and use, refer to the Taking
Thermocouple Temperature Measurements tutorial at ni.com/zone.
I believe this could be the cause of the fluctuations that you are seeing. With regard to calibration of your device the procedure to perform this is below.