Multifunction DAQ
Resistor Differential Measurement
I want to connect a floating voltage signal source to a NI-9202 in differential configuration. In the Getting-Started-Guide of the NI-9202 it is stated, that the negative lead has to be connected to COM through a 1 MΩ resistor.
However, the tutorial Grounding Considerations (https://www.ni.com/en/support/documentation/supplemental/06/grounding-considerations---intermediate-...) recommends 10kΩ<R<100kΩ.
Why is that difference and which way should I follow?
Is the  "AI GND"-Connection, shown in the tutorial, equal to the COM port of the NI-9202?
Thanks for your answers.
(966 Views)
Both is possible,
In short, what's best, depends on:
your Source, mainly output impedance and source impedances to EARTH/surroundings
your cable length and surrounding EMC
With two resistors the input impedance of your 9502 is 2R (instead of >10GOhm as stated in the spec), but bad vor high impedance sources (like ph probes)
If you have a low impedance source and longer cable, the 2 R solution help to minimize electric field influences.
Greetings from Germany
Henrik
LV since v3.1
(924 Views)
