Jerry,
1 - You will probably need to average several readings to help filter out any noise.
2 - It sounds like you need to scale the readings from your load cell. Based on part #, I'm guessing that your load cell output is rated for 3.0mV/V. That means that for every volt of excitation voltage, cell will output 3mV at fullscale. So in the case of a 100lb cell with 10V excitation, output is 30mV with 100lb load. The other part of this is that you need to account for any offset voltage at zero force applied to cell. You can measure this with no force applied to cell or you may have received a calibration sheet from manufacturer.
basic formula is (mV measured - zero offset)*fullscale force/fullscale output=force applied to cell